Hi
I have a sitrans probe a 12m ultrasonic transmitter for measuring levels and volumes of tanks,
I powered it up, but it stucks at the initialization screen.

Any idea how to solve this issue please?

Sayah
Top
  9/3/2009, 3:18 PM
Joined 3/21/2006
Last visit: 1/29/2015
Posts: 9931
Rating:
Rated: Outstanding [3.92 out of 5 / rated 1054 time(s)]. (1054) 
Rated: no ratings [0 out of 5 / rated 0 time(s)]. (0) 
Hello,

I suggest to contact with contact your local Siemens representative.blush

Reagrds
Top
  9/3/2009, 8:45 PM
Joined 9/25/2006
Last visit: 12/5/2014
Posts: 151
Rating:
Rated: Excellent [4.4 out of 5 / rated 5 time(s)]. (5) 
Rated: no ratings [0 out of 5 / rated 0 time(s)]. (0) 
Hi Sayah,

The Sitrans Probe is -as you know- a 2-wire instrument (and it's an excellent one!).
It needs some current to start up. Maybe your analog card does not provide enough power, or your loop resistance is too high. I have seen this many times.
By means of test, connect your Sitrans Probe directly to a 24V power supply (e.g. Sitop Power), and see if it does start up completely. When the start up phase is done, you will hear / feel the pulses on the sensor face.

Good Luck,

SonicRanger.
Top
  9/4/2009, 11:34 PM
Joined 9/3/2009
Last visit: 9/6/2009
Posts: 2
Rating:
Rated: no ratings [0 out of 5 / rated 0 time(s)]. (0) 
Rated: no ratings [0 out of 5 / rated 0 time(s)]. (0) 

Hi again,

I tried with independent PS and it works laugh.
However, I still need to connect it to my data acquesition system to enable reading in my system.
which is not possible as it's only 2wire sensor


I used a 250 Ohm resistor in series with the sensor and it WORKS.

Many thanks for your help blink

Cheers,
Sayah